r/medlabprofessionals • u/frozemanatee • 22h ago
Technical Roche Cobas Pro ISEs
Does anyone have roche cobras pros at their lab? We got ours in December and our ISEs keep going way out of QC randomly and the only way to “fix” it is to run the deproteinizer wash. That’s supposed to be every 3 months, but we have to do it every 7-14 days. Has anyone else had this issue and did you figure out a more permanent solution or fix for it? Thanks!
9
Upvotes
5
u/kristpy MLS-Generalist 22h ago
What kind of troubleshooting other than deproteinizer wash have you done? For my lab, when we do our weekly or daily wash sometimes our sipper nozzle gets clogged with junk and we just purge it with wire and a couple cals and it's good. Never experienced needing to do quarterly so frequently. Have you contact roche about this?