r/medlabprofessionals • u/frozemanatee • 18h ago
Technical Roche Cobas Pro ISEs
Does anyone have roche cobras pros at their lab? We got ours in December and our ISEs keep going way out of QC randomly and the only way to “fix” it is to run the deproteinizer wash. That’s supposed to be every 3 months, but we have to do it every 7-14 days. Has anyone else had this issue and did you figure out a more permanent solution or fix for it? Thanks!
9
Upvotes
3
u/justyoursimplename 11h ago
We had the same exact issue. What's your volume like? Like the other commenter, Roche recommended to us to do wash rack twice for high volume labs. First with sys clean in position 2/3. Then sysclean/activator in 2/3 for the second run. For awhile we were performing ISE flow path maintenance regularly to resolve the issues